Question (27521) | Cell Viability Assay Kit (Bioluminescent) (ab65314)

Go to datasheet (ab65314)

Question

Sure you suggest a luminometer but a beta counter as well in the protocol, for this reason I wrote you asking for more information about the Beta counter setup.

Answer

Thank you for your reply. I can confirm that conventional scintillation counters can not be used for measurements with this kit. However, when Beta Counter is used it should be programmed in the “out of coincidence” (or Luminescence mode) for measurement. So if they have this setting where they can program this, then the beta counter may be used. I hope this will be helpful to you. If you have any further questions, please do not hesitate to contact us.